Detection and Uses of Radioactivity — Question 9

Section 18-4 · Chemistry pages 3103–3104

Show solutionHide solution

Question 9

Use attenuation as a thickness measurement

A hypothetical narrow-beam gauge follows I/I0=e−μxI/I_0=e^{-\mu x} with μ=0.200mm−1\mu=0.200\ \mathrm{mm}^{-1}. Background-corrected transmitted intensity is 40.0 percent of the incident reference. Assume constant composition and negligible scattered photons at the detector.

Diagram shown in the original SVG worksheet.

  1. Find thickness x.

  2. Explain why a composition change could mimic a thickness change and why a reference reading is useful.

Question 9 - Solution

Use attenuation as a thickness measurement

Diagram shown in the original SVG worksheet.

(a) Reasoning and result

x=−ln⁡(0.400)/0.200=4.58145x=-\ln(0.400)/0.200=4.58145 mm.

(b) Analysis

Composition can change the attenuation coefficient, so a smaller transmitted fraction need not mean a thicker sheet. A reference monitors source output and detector response; it does not by itself remove uncertainty in material composition.

Check and interpretation

The exponential model applies to the stated narrow-beam idealization.