Gas Chromatography—Mass Spectrometry — Question 3

Section 15-14 · Organic Chemistry pages 3495–3496

Show solutionHide solution

Question 3

Challenge an identification based only on retention time

Both references fall inside the practical retention-time discrimination interval.

Diagram shown in the original SVG worksheet.

  1. Can retention time alone choose A over B?

  2. What additional GC–MS information can help?

  3. Why should an authentic standard be measured under matching conditions?

Question 3 - Solution

Challenge an identification based only on retention time

Diagram shown in the original SVG worksheet.

(a)

No. The difference is smaller than the stated repeatability.

(b)

Compare full mass spectra and diagnostic ions, ideally with authentic standards and additional separation evidence.

(c)

Retention and spectra can depend on column, temperature program and source conditions; unmatched conditions weaken the comparison.