Question 9
Quantify with a supplied internal-standard calibration
The response model has been validated over the relevant range.
Diagram shown in the original SVG worksheet.
Calculate the amount of X.
Why can an internal standard help with injection variability?
State one limitation if X and the standard behave differently during preparation.
Question 9 - Solution
Quantify with a supplied internal-standard calibration
Diagram shown in the original SVG worksheet.
(a)
nX = (0.90/1.50)(10.0) = 6.0 nmol.
(b)
A suitable internal standard experiences shared variation, so the area ratio can reduce injection-related error.
(c)
Different extraction losses, instability or matrix effects can remain; the internal standard and preparation method must be validated.