Question 10
Choose a protocol designed for quantitative carbon comparison
The aim is to compare carbon amounts rather than merely locate resonances.
Diagram shown in the original SVG worksheet.
Which protocol is designed to reduce the stated intensity biases?
Why is a sufficiently long recovery interval still needed?
Does the protocol name alone guarantee quantitative accuracy?
Question 10 - Solution
Choose a protocol designed for quantitative carbon comparison
Diagram shown in the original SVG worksheet.
(a)
B is designed to reduce relaxation and NOE-related biases.
(b)
Different carbon sites must recover sufficiently to give comparable responses under the chosen pulse scheme.
(c)
No. Calibration, signal-to-noise, overlap, pulse settings and recovery must be validated for the sample and required precision.